SALUDOS. emolina2122@hotmail.com Error numbers If either message listed below is displayed, first power OFF, then ON again. If the message reappears, note the error number displayed in the high SpO2 alarm limit area and refer to the chart below for guidance. Error Number Cause Action 1–24, 26, 27 Faulty sensor. Replace the sensor. 25 Incompatible sensor. Use only OxyTip+ sensors. 28, 29 System board failure. Replace the System board. 27, 79, 81, 82 Software malfunction. Install a software upgrade. 83 Battery failure. Replace the battery pack. 89, 91, 117,118 Serial communication error when printing or downloading. Check cable and connections to printer/PC. Cycle power and retry. 142 Trend memory malfunction. Use the Trend Download software to erase trend memory. |
||